Full-View 3D Photoacoustic and Ultrasound Breast Tomography Published …
Full-View 3D Photoacoustic and Ultrasound Breast Tomography Published in Photoacoustics
July 22, 2026
A European research team has published results on PAM3, an integrated 3D photoacoustic and ultrasound breast imaging system, in the journal Photoacoustics. The study, featured on the cover of Volume 50, details how the system combines light and sound to improve non-invasive imaging in dense breast tissue.
The paper, titled “Hybrid three-dimensional full-view multi-wavelength photoacoustic and ultrasound breast tomography”, presents a dual-modality system designed to solve a common problem in photoacoustic imaging: sound travelling at different speeds through different types of biological tissue, which distorts the final image.
How the Technology Works
Photoacoustic imaging works by combining light absorption with ultrasound detection. Short pulses of laser light illuminate the tissue, where natural absorbers, mainly hemoglobin in blood vessels, absorb the energy and expand slightly. This expansion creates weak ultrasound waves that travel back to the surface, where acoustic sensors detect them to build a 3D image of blood vessel networks.
Because tumors need new blood vessels (angiogenesis) to grow, mapping blood vessel density and blood oxygen levels gives doctors useful diagnostic details without using ionizing radiation or contrast agents.
However, breast tissue is a mix of fat, glandular tissue, and lesions, and sound travels through each at slightly different speeds. If an imaging system assumes a single average sound speed, the reconstructed image becomes blurry. The PAM3 system solves this by first using ultrasound to measure the exact sound speed at every point in the breast. It then uses this 3D sound map to correct the photoacoustic image, keeping blood vessel networks sharp and accurate.
In vivo photoacoustic and sound-speed breast image made with the PAM3 system.
Sound-speed-compensated projections of the right breast of a healthy volunteer. From Photoacoustics 50 (2026), 100847.
Optical Excitation: EKSPLA PhotoSonus Laser Integration
To deliver the required multi-wavelength optical pulses, the PAM3 setup incorporated an EKSPLA PhotoSonus series tunable wavelength laser as its photoacoustic excitation source. The source comprises two identical units, each consisting of a Q-switched Nd:YAG pump laser, a DKDP second harmonic generator, and a BBO optical parametric oscillator producing 5 ns pulses; their outputs are combined into a single beam coupled to the fiber bundle.
Designed specifically for biomedical and photoacoustic imaging applications, the PhotoSonus series provides a combined output energy from 450 mJ at 680 nm to 230 mJ at 1060 nm, roughly three times the energy reported for lasers in the same class, and 1.5 times per laser-OPO unit. Its fully motorized and computer-controlled architecture allowed for straightforward hardware integration into the PAM3 control software. Operational features used during the study included synchronized external trigger inputs and outputs, internal energy monitoring, a motorized optical attenuator, and automated switching between OPO signal and idler wavelengths.
Performance Results Reported in the Study
- Spatial Resolution: ~0.7 mm isotropic in 3D (786 μm in x, 775 μm in y, 693 μm in z), improving to 426 μm in z with 100 reconstruction iterations, comparable to the in-plane resolution of clinical breast MRI.
- Imaging Depth: Captures small blood vessels up to 48 mm deep inside tissue.
- Dual Data: Provides two complementary views at once: blood vessel distribution (optical) and tissue sound speed (ultrasound).
The project was carried out by a multi-institutional European collaboration of academic, industrial, and clinical partners, including the University of Twente (TechMed Centre), Center for Physical Sciences and Technology (FTMC), Centrum Wiskunde & Informatica (CWI), University College London (UCL), Brno University of Technology, P.A. Imaging R&D B.V., Imasonic SAS, Leiden University Medical Center (LUMC), Ziekenhuisgroep Twente (ZGT), and Medisch Spectrum Twente (MST).
전방향 3D 광음향 및 초음파 유방 단층촬영 기술, 국제 학술지 Photoacoustics에 게재
2026년 7월 22일
유럽 공동 연구팀이 통합형 3D 광음향 및 초음파 유방 영상 시스템인 PAM3의 연구 결과를 국제 학술지 Photoacoustics에 발표했습니다. 이번 연구는 해당 학술지 제50권의 표지 논문으로 선정되었으며, 빛과 소리를 결합하여 치밀 유방 조직에서도 보다 정밀한 비침습적 영상 구현이 가능한 기술을 소개합니다.
논문 제목은 **「Hybrid three-dimensional full-view multi-wavelength photoacoustic and ultrasound breast tomography(하이브리드 3차원 전방향 다중 파장 광음향 및 초음파 유방 단층촬영)」**이며, 광음향 영상 기술의 일반적인 문제점 중 하나인 생체 조직별 음속 차이로 인해 발생하는 영상 왜곡을 해결하기 위해 개발된 이중 영상 방식의 시스템을 제시합니다.
기술 작동 원리 (How the Technology Works)
광음향 영상(Photoacoustic Imaging)은 빛의 흡수와 초음파 검출 기술을 결합한 영상 기법입니다.
짧은 레이저 펄스를 조직에 조사하면 혈관 내 헤모글로빈을 비롯한 생체 내 광흡수 물질이 빛 에너지를 흡수하고 미세하게 팽창합니다. 이 과정에서 약한 초음파가 발생하며, 조직 표면의 음향 센서가 이를 감지하여 혈관 네트워크의 3차원 영상을 재구성합니다.
종양은 성장 과정에서 새로운 혈관을 형성하는 혈관신생(Angiogenesis)이 필요합니다. 따라서 혈관 밀도와 혈중 산소 농도를 영상화하면 의료진에게 유용한 진단 정보를 제공할 수 있습니다. 또한 방사선이나 조영제를 사용하지 않는 비침습적 영상 진단이 가능하다는 장점이 있습니다.
그러나 유방 조직은 지방, 유선 조직, 병변 등 다양한 조직으로 구성되어 있으며, 각 조직을 통과하는 초음파의 속도는 서로 다릅니다. 영상 시스템이 평균 음속 하나만을 기준으로 영상을 재구성할 경우, 이러한 차이로 인해 영상이 흐려지거나 왜곡될 수 있습니다.
PAM3 시스템은 초음파를 이용해 유방 내부 각 지점의 실제 음속을 먼저 측정하고, 이를 3차원 음속 지도로 구현하여 광음향 영상의 재구성 과정에 반영합니다. 이를 통해 조직 내 혈관 네트워크를 더욱 선명하고 정확하게 영상화할 수 있습니다.
건강한 지원자의 오른쪽 유방을 대상으로 PAM3 시스템으로 촬영한 생체 내 광음향 영상 및 음속 영상. Photoacoustics 제50권(2026), 논문번호 100847.광학적 여기: EKSPLA PhotoSonus 레이저 통합 (Optical Excitation: EKSPLA PhotoSonus Laser Integration)
다중 파장의 광 펄스를 구현하기 위해 PAM3 시스템에는 광음향 여기 광원으로 EKSPLA PhotoSonus 시리즈 가변 파장 레이저가 적용되었습니다.
광원은 동일한 두 개의 레이저 유닛으로 구성되며, 각 유닛은 Q-스위칭 Nd:YAG 펌프 레이저, DKDP 제2고조파 발생기(Second Harmonic Generator), 그리고 5ns 펄스를 생성하는 BBO 광파라메트릭 발진기(Optical Parametric Oscillator, OPO)로 이루어져 있습니다.
두 유닛에서 출력된 빔은 하나의 빔으로 결합되어 광섬유 번들(Fiber Bundle)을 통해 영상 시스템으로 전달됩니다.
생체의료 및 광음향 영상 응용 분야를 위해 특별히 설계된 PhotoSonus 시리즈는 다음과 같은 높은 출력 에너지를 제공합니다.
680nm 파장에서 450mJ
1060nm 파장에서 230mJ
이는 동일 등급의 레이저와 비교해 약 3배 높은 출력 에너지이며, 레이저-OPO 개별 유닛 기준으로도 약 1.5배 높은 에너지 성능에 해당합니다.
또한 완전 자동화된 모터 구동 방식과 컴퓨터 제어 아키텍처를 갖추고 있어 PAM3 제어 소프트웨어와의 하드웨어 통합이 용이했습니다.
이번 연구에서는 다음과 같은 주요 기능이 활용되었습니다.
외부 트리거 입력 및 출력 동기화 기능
내부 레이저 출력 에너지 모니터링 기능
모터 구동식 광 감쇠기(Optical Attenuator)
OPO 시그널(Signal) 및 아이들러(Idler) 파장 간 자동 전환 기능
이러한 기능은 정밀한 다중 파장 광음향 영상 구현을 위한 안정적인 레이저 제어와 시스템 통합을 지원합니다.
연구에서 보고된 주요 성능 (Performance Results Reported in the Study)
- 공간 분해능(Spatial Resolution): 3차원 영상에서 약 0.7mm의 등방성 분해능을 구현했습니다. 세부적으로 x축 786μm, y축 775μm, z축 693μm의 분해능을 보였으며, 100회의 영상 재구성 반복을 적용했을 때 z축 분해능은 426μm까지 향상되었습니다. 이는 임상용 유방 MRI의 평면 내 분해능에 필적하는 수준입니다.
- 영상 촬영 깊이(Imaging Depth): 조직 내부 최대 48mm 깊이까지 미세 혈관을 영상화할 수 있습니다.
- 이중 영상 데이터(Dual Data): 혈관 분포를 보여주는 광학 영상과 조직의 음속 정보를 제공하는 초음파 영상을 동시에 확보하여 상호 보완적인 진단 정보를 제공합니다.
공동 연구 참여 기관 (Research Collaboration)
이번 프로젝트는 학계, 산업계 및 임상 분야의 여러 기관이 참여한 유럽 다기관 공동 연구로 진행되었습니다.
참여 기관은 다음과 같습니다.
University of Twente (TechMed Centre)
Center for Physical Sciences and Technology (FTMC)
Centrum Wiskunde & Informatica (CWI)
University College London (UCL)
Brno University of Technology
P.A. Imaging R&D B.V.
Imasonic SAS
Leiden University Medical Center (LUMC)
Ziekenhuisgroep Twente (ZGT)
Medisch Spectrum Twente (MST)
관련 논문 (Related Publications)